As I mentioned earlier, I was permitted to work in the director's laboratory, which was far better equipped than our own. Unlike many other research institute directors, ours was open to discussing scientific topics and would sometimes even heed the opinions and arguments of his staff. He urgently needed a method for determining the moisture content of urotropine without relying on expensive equipment. It took some time, but I eventually managed to explain to him that the standard drying method was unsuitable - at least in its classical form. However, the clients insisted on their requirements, and the director was reluctant to refuse them. They had even paid for an automatic Karl Fischer titrator(!), but delivery of this costly instrument would take months, whereas the results were needed immediately.

In addition to standard analytical and technical balances, desiccators, weighing bottles, and a drying oven, the laboratory possessed a moisture analyzer that operated on the drying principle. Essentially, this single device replaced all the aforementioned laboratory equipment. It significantly accelerated and simplified moisture analysis, although the moisture analyzer's accuracy was lower. The instrument consisted of an analytical balance topped by two halogen lamps designed to heat and dry the sample. Both the temperature and the heating duration could be easily adjusted. The device provided an immediate readout of the percentage of mass loss, along with a graph showing mass loss over time.

The moisture analyzer had been purchased recently, and there wasn't a single chemist in the director's laboratory capable of setting it up. Consequently, the task of getting the instrument running fell to me and to a chemistry student who was working in the director's lab on a probationary basis. For lack of a better option, I suggested using a piece of Ukrainian rye bread as the test sample. The moisture content of bread can vary significantly, and for this particular sample, it was unknown beforehand. However, we first needed to verify that the moisture analyzer was actually working, so we placed the bread inside and set it to heat at 105°C for one hour.

Ukrainian Rye Bread
Ukrainian Rye Bread


The results ranged from 31% to 39%. The bread was uneven in terms of moisture content - its outer layers had partially dried out. The student was unaware that the initial sample needed not only to be cut into small pieces but also thoroughly homogenized. I noticed this before the analysis began but chose not to intervene; the exact moisture content of the Ukrainian bread wasn't so important. The main thing was that the instrument was functioning and the readings were plausible.

Since the smell of toasted bread wafted through the entire laboratory after the moisture determination, I nicknamed the analyzer the "bread maker." None of the other staff used this term, but whenever I said "bread maker," everyone knew exactly which instrument I was referring to.

Bread Maker


I flatly refused to heat hexamine in that moisture analyzer; the substance would sublime and then condense on the instrument's electronics, potentially ruining this wonderful piece of equipment. The director agreed with me and did not press the issue - especially since, in addition to hexamine, we needed to determine the moisture content of another material.

This substance was stable at 105°C and non-hygroscopic; it did not evaporate even at much higher temperatures. However, it had its own unpleasant characteristics - more serious ones than those of hexamine. Furthermore, when I consulted the regulatory document, I found neither a specified limit for moisture and volatile matter content nor a reference to a recommended determination method. However, I decided that a "bread maker" would serve the purpose.

The moisture analyzer did not let us down; with its help, we determined the moisture content of the samples quickly and easily. The first analysis showed a water content of 8.7%, the next 11.2%... The results ranged from 2.7% to 14.4%. Gravimetric analysis performed using the classical method - involving an analytical balance, a drying oven, and a desiccator-yielded a similarly wide spread of results.

The reason turned out to be the heterogeneity of the sample submitted for analysis. Once I homogenized the sample, which weighed nearly 1 kg, the results of parallel analyses became much more consistent, with an average value of 13.7%. The clients claimed they had homogenized the sample before handing it over to us. They might have been telling the truth, or perhaps not. However, the sample container was not hermetically sealed. Furthermore, it later emerged that the sample tended to stratify over time, with water settling at the bottom. Consequently, the moisture content could be 3% at the top of the sample and 18% near the bottom of the container.

A clear analogy is a bucket of dry river sand to which a little water has been added. The water gradually seeps downward; consequently, the sand at the top soon becomes dry again, while the sand near the bottom of the bucket becomes saturated with liquid water. A similar phenomenon can often be observed at the beach. Even with river sand, things are not quite so simple.



'Хлебопечка' (весовой анализатор влаги) - Часть 9
Как я уже упоминал, мне разрешили поработать в лаборатории директорского отдела, которая была оснащена значительно лучше, чем наша лаборатория. В отличие от многих других директоров исследовательских институтов, с директором нашего института вполне можно было обсуждать научные темы, иногда он даже прислушивался к мнению и аргументам сотрудников. Директору был очень нужен метод определения влаги в уротропине без использования дорогого оборудования. Далеко не сразу, но мне удалось ему объяснить, что метод высушивания здесь неприменим, по крайней мере, в его классическом варианте. Однако, заказчики настаивали на своем, и директор не хотел им отказывать. Они даже оплатили покупку автоматического титратора по Карлу Фишеру(!), но поставка этого дорогого прибора требовала несколько месяцев, а результаты были нужны сейчас.

Кроме обычных аналитических и технических весов, эксикаторов, бюксов и сушильного шкафа, в лаборатории был также анализатор влаги, работа которого основана на методе высушивания. Один этот прибор, по сути, заменял все перечисленное лабораторное оборудование. Анализ содержания воды ускорялся и значительно упрощался, хотя точность влагомера была ниже. Прибор представлял собой аналитические весы, над которыми располагались две галогенные лампы, предназначенные для нагрева и высушивания исследуемого образца. Температуру и время нагревания образца можно было легко регулировать. Прибор сразу выдавал процент потери массы и график зависимости потери массы от времени.

Влагомер недавно купили и в директорской лаборатории не было ни одного химика, чтобы запустить его в работу. Поэтому запускать прибор пришлось мне вместе со студентом химического факультета, который работал в директорской лаборатории, находясь на испытательном сроке. За неимением лучшего, качестве тестового образца я предложил использовать... кусок украинского ржаного хлеба. Содержание влаги в хлебе может сильно варьироваться, причем для данного образца оно было неизвестно заранее. Однако, для начала нам нужно было проверить, работает ли влагомер вообще, поэтому мы поместили внутрь хлеб и включили нагрев 105°С на 1 час.

Ukrainian Rye Bread
Ukrainian Rye Bread


Полученные результаты колебались от 31 до 39%. Хлеб был неоднородный по содержанию влаги - его внешние слои частично подсохли. Студент не знал, что исходный образец нужно было не только порезать на мелкие кусочки, но и тщательно гомогенизировать. Я заметил это еще до анализа, но вмешиваться не стал - точное содержание влаги в украинском хлебе было не так важно. Главное, что прибор работал и его показания были правдоподобны.

Поскольку после определения влаги по всей лаборатории распространился запах жаренного хлеба, я назвал влагомер "хлебопечка". Кроме меня никто из сотрудников этот термин не употреблял, но, когда я говорил "хлебопечка", все понимали, о каком приборе идет речь.

Bread Maker


Нагревать гексамин в этом влагомере я категорически отказался - вещество будет сублимироваться, а затем конденсируется на электронике прибора. Хороший прибор мог выйти из строя. Директор со мной согласился и настаивать не стал. Тем более, что кроме уротропина было необходимо определить влагу в другом материале.

Вещество это было стабильно при 105°С и негигроскопичное. Даже при гораздо более высоких температурах оно не испарялось. Но были и у него свои неприятные особенности - более серьезные, чем у гексамина. Плюс к этому, когда я посмотрел нормативный документ, то не обнаружил в нем ни нормы содержания влаги и летучих веществ в данном веществе, ни ссылку на рекомендованный метод определения. Однако, решил, что "хлебопечка" для этих целей подойдет.

И весовой анализатор влаги не подвел - с его помощью мы определили влажность образцов легко и быстро. Первый анализ показал содержание воды 8.7%, следующий - 11.2%... Результаты варьировались от 2.7 до 14.4%. Гравиметрический анализ, проведенный классическим методом - с помощью аналитических весов, сушильного шкафа и эксикатора - давал аналогичный разброс результатов.

Причиной оказалась неоднородность образца, который нам принесли на анализ. Когда я усреднил образец (который весил почти 1 кг), результаты параллельных анализов стали гораздо ближе друг к другу, а среднее значение составляло 13.7%. Заказчики утверждали, что они усреднили образец перед тем, как отдать его нам на анализ. Возможно, они сказали правду, возможно - нет. Однако, тара с образцом была негерметично закрыта. Кроме того, позже оказалось, что образец со временем расслаивается - вода опускается вниз. В результате, сверху влажность образца могла быть 3%, а возле дна контейнера - 18%.

Наглядная аналогия - сухой речной песок в ведре, в который добавили немного воды. Вода постепенно просочится вниз, в результате сверху песок скоро снова будет сухим, зато возле дна ведра он станет насыщенным жидкой водой. Подобное явление часто можно наблюдать на пляже. Даже с речным песком не все так просто.


Download the Video (83 Mb, .avi )

'Bread Maker' (Gravimetric Moisture Analyzer)
'Bread Maker' (Gravimetric Moisture Analyzer)

'Bread Maker' (Gravimetric Moisture Analyzer)

'Bread Maker' (Gravimetric Moisture Analyzer)

'Bread Maker' (Gravimetric Moisture Analyzer)

'Bread Maker' (Gravimetric Moisture Analyzer)

'Bread Maker' (Gravimetric Moisture Analyzer)

'Bread Maker' (Gravimetric Moisture Analyzer)

'Bread Maker' (Gravimetric Moisture Analyzer)

'Bread Maker' (Gravimetric Moisture Analyzer)

'Bread Maker' (Gravimetric Moisture Analyzer)

'Bread Maker' (Gravimetric Moisture Analyzer)

'Bread Maker' (Gravimetric Moisture Analyzer)

'Bread Maker' (Gravimetric Moisture Analyzer)

'Bread Maker' (Gravimetric Moisture Analyzer)

'Bread Maker' (Gravimetric Moisture Analyzer)

'Bread Maker' (Gravimetric Moisture Analyzer)




Having noticed a mistake in the text, allocate it and press Ctrl-Enter


Interesting Thought Occurred to Me - Part 10
Возникла интересная мысль - Часть 10
Over the years, I have performed numerous gravimetric analyses to determine the water content of a wide variety of samples. Since the substances and technical products varied, their dehydration temperatures differed significantly, as did the specific procedures employed. For instance, ground sausage or fatty fish were mixed with dry quartz sand and heated in a drying oven at 105°C for one hour. Butter, on the other hand, was dehydrated in aluminum weighing bottles on an electric hot plate rather than in an oven. While working on my dissertation, I heated crystalline hydrates of divalent metal phosphates in a muffle furnace at 800°C for one hour.

However, all the analyses mentioned above share a common feature. Let us consider one of the aforementioned phosphates - cobalt phosphate, Co3(PO4)2·8H2O, for example. If the heating time in the muffle furnace had been different - say, fifty minutes or three hours instead of one hour - it would not have fundamentally affected the results of the chemical analysis. Nor would it have mattered if the calcination temperature had been 913°C instead of 800°C.

The crucial requirement is that all the water - hygroscopic, constitutional, and water of crystallization - be removed during heating. There is a term for this: "heating to constant mass." Once constant mass has been reached, extending the heating time makes no difference because all the water has already been driven off. In most cases, I heated samples for one hour, as prescribed by the analytical procedures. I sometimes wondered how much this duration could be reduced and even conducted preliminary experiments with various types of samples. However, shortening the heating time offered no significant practical advantage. While the samples were heating in the drying oven, I could carry out other work alongside the determination of their water content.

The situation with hexamine was different. Initially, both water and the hexamine itself would evaporate from the sample simultaneously; subsequently, only the hexamine would continue to evaporate. Increasing the heating time or temperature resulted in greater mass loss, causing the analysis results to deviate increasingly from the sample's actual water content.

If one could pinpoint the exact moment when water evaporation ceased and stop heating the sample at that precise moment, the analysis would yield the most accurate results - or, more precisely, the least inaccurate ones.

A physicist colleague proposed an idea. Since water evaporation and urotropine sublimation occur at different rates, one could plot a kinetic curve showing the sample's mass loss as a function of heating time. Theoretically, this would make it possible not only to determine when all the water had evaporated but also to calculate the mass of urotropine that had sublimed along with it. This is a standard type of task in thermogravimetric analysis.

In practice, however, neither our laboratory nor even the director's laboratory possessed the equipment required for thermogravimetric analysis. The curve would have to be plotted manually: heat the sample for a set period, cool it in a desiccator, and weigh it; then heat it again, cool it, and weigh it once more, and so on. The accuracy of such a determination would be unsatisfactory, and the process would require a great deal of time and effort - especially given that the water content of the hexamine was merely a fraction of a percent.

I had actually conceived the same idea before the physicist mentioned it, but I realized that it wouldn't work: it required a thermogravimetric analyzer. And even then, the data would only be approximate.

As time went by, I suddenly realized that the laboratory already had an instrument that could serve this purpose - one that had been sitting right under my nose. In fact, I had recently used it. It was called a "gravimetric moisture analyzer."

In addition to measuring total mass loss, the analyzer also tracks how mass loss changes over time. When I worked with the different substance - one that neither evaporates nor decomposes at 105°C (see the previous part of the article) - that curve was of merely aesthetic interest. I was concerned with the moisture content of the samples, not the rate at which they lost mass during the analysis.

However, I noticed that just five minutes was sufficient to remove the water from 1-2 g samples; there was no need to heat them for a full hour. I then adjusted the instrument's settings to a heating time of 25 minutes, but even that proved excessive. Much later, I read the user manual, which recommended a method for determining the analysis duration: first, establish how long it takes to remove all the moisture, using the curve mentioned above or the instrument's numerical readings, and then add 15-20% to that time.

It occurred to me that if roughly five minutes was enough to dry the substance I had been working with, the same would almost certainly be true of hexamethylenetetramine. Heating hexamine for five minutes made sense. Although this would not prevent sublimation entirely, it would significantly reduce the amount of hexamine lost, since five minutes is twelve times shorter than one hour.

The only thing that stopped me from testing this idea was my reluctance to ruin the moisture analyzer.



Возникла интересная мысль - Часть 10
За много лет я провел большое количество гравиметрических анализов, определяя содержания воды в разнообразных образцах. Поскольку вещества или технические продукты были разные, их температура дегидратации могла существенно отличаться, обличались и особенности методик. Например, перемолотую колбасу на мясорубке или жирную рыбу смешивали с сухим кварцевым песком и нагревали в сушильном шкафу при 105°С на протяжении одного часа. Зато сливочное масло обезвоживали не в сушильном шкафу, а в алюминиевых бюксах на электрической плитке. Когда я работал над диссертацией, нагревал кристаллогидраты фосфатов двухвалентных металлов в муфельной печи при температуре 800°С один час.

Однако, все перечисленные анализы имеют одну общую особенность. Рассмотрим упомянутые фосфаты, например, фосфат кобальта Co3(PO4)2·8H2O. Если бы время нагревания в муфельной печи было другим, скажем, не один час, а пятьдесят минут или три часа, то это не имело бы принципиального значения для химического анализа. Если бы температура прокаливания была не 800°С, а 913°С, это тоже было бы неважно.

Важно, чтобы в процессе нагревания вся вода испарилась - и гигроскопическая, и кристаллогидратная, и конституционная. Для этого существует термин "нагреть до постоянной массы". Дальнейшее увеличение времени нагревания уже роли не играет, поскольку вся вода уже улетучилась. В большинстве случаев я нагревал образцы один час - этого требовали методики анализов. Я иногда задумывался, насколько можно было уменьшить это время. Даже проводил пробные эксперименты с разными типами образцов. Однако сокращение времени нагревания не имело большого практического значения. Пока образцы нагревались в сушильном шкафу, можно было выполнять другую работу - одновременно с определением содержания воды.

В случае с гексамином все было иначе. Сначала из образца одновременно испарялись вода и сам гексамин, а потом - только гексамин. Увеличение продолжительности и температуры нагревания приводит к увеличению потери массы, следовательно, результаты анализа становятся все более далекими от истинного содержания воды в образце.

Если бы удалось точно определить момент, когда испарение воды закончилось, и прекратить нагревание образца именно в этот момент, то результаты анализа были бы наиболее корректными (точнее - наименее некорректными).

Коллега-физик предложил идею. Поскольку испарение воды и сублимация уротропина происходят с разными скоростями, можно построить кинетическую кривую зависимости потери массы образца от времени нагревания. Теоретически это позволило бы не только определить момент, когда вся вода испариться, но и рассчитать массу уротропина, который испарился одновременно с водой. Для термогравиметрического анализа это стандартная задача.

На практике же, ни наша лаборатория, ни даже директорская лаборатория не имели приборов, необходимых для термогравиметрического анализа - кривую пришлось бы строить вручную, каждый раз нагревая образец заданное время, охлаждая в эксикаторе и взвешивая. Потом - повторный нагрев и т.д. Точность такого определения была бы неудовлетворительной, а времени и труда понадобилось бы много. Тем более, что содержание воды в гексамине составляло всего десятые доли процента.

Описанная идея пришла мне в голову тоже - еще раньше, чем ее высказал физик, но я понял, что ничего не получится - нужен прибор для проведения термогравиметрического анализа. И даже в этом случае данные будут только ориентировочными.

Прошло время, и я неожиданно я осознал, что такой прибор в лаборатории есть, и он стоит буквально под носом. Более того, я на этот приборе недавно работал, и называется он "гравиметрический анализатор влажности".

Кроме потери массы, анализатор также дает зависимость потери массы от времени. Когда я работал с другим веществом, которое не испаряется и не разлагается при 105°С (см. предыдущую часть статьи), данная кривая имела разве что эстетическое значение. Меня интересовал процент влаги в образцах, а не скорость ее испарения в процессе анализа.

Однако, я заметил, что для удаления воды из образцов массой 1-2 грамма хватало буквально пяти минут - не было потребности нагревать их целый час. Тогда я установил в настройках прибора время нагревания в 25 минут, но и оно было излишним. Намного позже я прочитал руководство пользователя прибора, в котором рекомендуется способ вычисления продолжительности анализа. Сначала следует установить время, необходимое для полного удаления влаги из образца, - исходя из упомянутой кривой или из числовых показаний прибора, затем нужно добавить к этому времени 15-20%.

Возникла мысль: поскольку для высушивания данного вещества хватает примерно 5 минут, то почти наверняка это верно и для гексаметилентетрамина. Целесообразно нагреть гексамин 5 минут - это не предотвратит его сублимацию, но значительно уменьшит данный нежелательный процесс, ведь 5 минут - это в 12 раз меньше, чем один час.

Единственное, что меня удержало от того, чтобы проверить данную идею, - нежелание угробить анализатор влаги.


Gravimetric Moisture Analyzer
Gravimetric Moisture Analyzer

Gravimetric Moisture Analyzer

Gravimetric Moisture Analyzer

Gravimetric Moisture Analyzer

Gravimetric Moisture Analyzer



[ Qualitative and Quantitative Chemical Analysis ] [ Chemical laboratory. Glassware. Reagents. Laboratory operations ]

Gravimetric (weight) analysis - Discussion on the forum
Qualitative analysis (qualitative reactions, separation...) - Discussion on the forum
Chemical photographs and laboratory videos - Discussion on the forum
[Submit a Comment / Error Message - Отправить Комментарий / Сообщение об ошибке]